Leaf photosynthesis MultispeQ v2 Protocol Run Time
Technical Support and Questions
Habib-ur-Rehman Athar
Mar 2019
Hello, I would like to inquire about protocol run time. I am using MultispeQ v 2.0; Firmware 2.0005 I checked following protocol runtime. Leaf Photosynthesis MultispeQ v 1 50 Second Lead Photosynthesis MultispeQ v 2 50 Seconds RIDES Protocol 1 min 20 seconds
Is it fine or we need to adjust our instrument, Open and Close Clamp position. On website, protocol runtime is <15 Seconds Measures the same photosynthesis-related parameters in <15 seconds as 'Leaf Photosynthesis MutlispeQ v1.0' plus additional PS1 and P700 measurements.
Sebastian
Mar 2019
It sounds like an issue with the calibration of the open and close position. I would calibrate the positions for two reasons:
- When clamping the leaf and waiting for an extended period of time before running the protocol, you might perturb steady state photosynthesis, which might causes a less accurate measurement of the photosynthetic parameters
- You save a lot of time when doing large scale experiments
Habib-ur-Rehman Athar
Mar 2019
Hello Sebastian Would you please send me calibration of the open and close position. Either we need to do as follow or something else 1- Connect the device with Desktop application 2- Click on setting Tab..... A pop up window open with three Tab 3- Click on Instrument Settings Tab (central One) 4- Make the Clamp Open .... 5- Press the button Set ---- Clamp Position Open 6 - Close the Clamp while placing leaf.... 7- Press the button Set ... Clamp Position
Sebastian
Mar 2019
Please follow these instructions as discussed here.
- Open the Desktop App and connect your Instrument
- Select View->Console from the menu to open the console
- Use the command
set_open_closed_positions
followed by enter - You get asked to set the open position = 4mm and closed position = 2mm
- Use
+
followed by enter after each position.
~Sebastian